Steven de Boer wrote: ↑Mon Dec 18, 2023 9:09 pm
We are a new user and are currently treating without beam control. We are noticing that the displayed deltas will be out of tolerance only when some cameras are blocked by the rotating gantry. Once the camera is no longer blocked the deltas are again within acceptable range. The duration of out of tolerance deltas is many seconds. How are people treating VMAT patients with active beam control?
We have been treating SRS (VMAT, HyperArc) with AlignRT with active beam control without any camera obstruction issues. A few things come to mind while reading your post. First, I would check to make sure the monthly physics camera calibration looks good (also good to check the daily for any drifts). Secondly, I would check the ROI you are using with the plan, make sure it's the right size and location for the treatment you are doing. We (and other people) have evaluated out AlignRT system to have submillimeter accuracy even with camera obstruction. At our clinic we see a mac of 0.5mm deviation with camera obstruction, which is well under out 1mm tolerance.
I'd be happy to look at your setup, feel free to contact me
Adi