I have two questions:
(1) Is everyone simulating brain SRTs with chin neutral or down? If not, why not?
(2) What is your process when RTDs exceed 1 mm after the couch has been rotated, particularly to 90 or 270? To exclude patient motion, we return the couch to zero and if the RTDs are negligible, we conclude that the patient has not moved. Then, we rotate the couch back to 90/270 and acquire an AP or PA MV image at the couch angle. At this point we must make a decision: is the RTD real or artifact of the VisionRT system at extreme couch angles? Do we shift or not?
Thank you,
Thanks,
(1) Is everyone simulating brain SRTs with chin neutral or down? If not, why not?
(2) What is your process when RTDs exceed 1 mm after the couch has been rotated, particularly to 90 or 270? To exclude patient motion, we return the couch to zero and if the RTDs are negligible, we conclude that the patient has not moved. Then, we rotate the couch back to 90/270 and acquire an AP or PA MV image at the couch angle. At this point we must make a decision: is the RTD real or artifact of the VisionRT system at extreme couch angles? Do we shift or not?
Thank you,
Thanks,