Breast treatment - using deformation tool

Posted: Mon Feb 03, 2025 5:41 pm
by julie_wood
We are currently having Align RT installed.

It seems most departments still perform a field light coverage check for breast treatments when using SGRT for set up. Has anyone replaced this with a check of the deformation tool? If the breast hasn't changed shape then coverage should be fine?

regards
Julie

Re: Breast treatment - using deformation tool

Posted: Wed Feb 19, 2025 10:05 am
by joshua_naylor
we stopped doing SSD check when introduced SGRT. Our procedures still have the rads do field light check just for splash (flash?! - external surface). The feeling is that it doesn't add much with SGRT, and we also do an MV-kV pair at post oblique angle so the MV would show any issue too.