Forums General AlignRT use with Elekta Linacs Reply To: AlignRT use with Elekta Linacs


ANANDAKRISHNAN K
Participant

Hi Jose, Blip of RTD’s at particular gantry angles can be fixed (even with hexapod frame) by increasing the ROI(Little away from Treatment iso) where Pod 3’s projections are not block on patient skin. Thank you Ben Allen for your nice suggestion to increase the ROI. As i said in my previous comment appropriate ROI will fix this blip of RTD’s. I think we need to accept the benefit of kV detector at 60cm from iso, through which we can acquire CBCT even at couch lateral 10cm. Once again Thank you all.